I have had almost only bad experience with TFA, that´s why I didn´t mention it, but what happened? What compositions did you use?

Does the tailing occur at all loadings or only under your prep conditions?
the tailing happens even at analytical level.
Going only by what has been written in this thread, I am not sure you have investigated the loading issue? If the application was already tailing at the analytical level and you performed a standard scale-up (i.e. multiply by the ratio of column volumes) then surely we would expect the prep application to exhibit tailing.

Can you clarify for us whether you investigated loading on the analytical column before scale-up?

You have stated the application is normal phase, but what is the column?
17 posts Page 2 of 2

Who is online

In total there are 10 users online :: 1 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry