I am looking for advice on the best way to regenerate a C18 PepMap100 Column. Does anyone have experience with this?
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
The PepMap stationary phase is a conventional monomerically bonded C18. (Carefully characterized and controlled for peptide mapping.) You would clean it in the usual ways. To remove hydrophobic contaminants, wash with 50:50 MeCN:H2O then acetone then MeCl2 then acetone then 50:50 MeCN:H2O.
If the problem is loss of bonded phase there is no cure. If there is a void, it is not practical to top off a capillary column. You have to replace the column.
If the problem is loss of bonded phase there is no cure. If there is a void, it is not practical to top off a capillary column. You have to replace the column.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 3
- Joined: Fri Sep 14, 2007 12:03 am
Thanks Mark for your input.
I am not sure why the column has lost capacity. It is a 300um capillary column that we are using for 2-D chromatography of peptides. Can the column retain or be effected by ionic species, such as the ammonium acetate we're using in the 1st dimension of our separation? If so, is there a cleanup that would clean the column of these ionic species?
Thanks
I am not sure why the column has lost capacity. It is a 300um capillary column that we are using for 2-D chromatography of peptides. Can the column retain or be effected by ionic species, such as the ammonium acetate we're using in the 1st dimension of our separation? If so, is there a cleanup that would clean the column of these ionic species?
Thanks
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Exposure to ionic species will not cause contamination of the column. What is the pH of the ammonium acetate from the 1st dimension? How much organic in the 1st dimension mobile phase? How large an injection? Has the void peak shifted?
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 3
- Joined: Fri Sep 14, 2007 12:03 am
Here's the info. you requested. Thanks for the help.
The pH is 7.2 of the 1000mM ammonium acetate solution. We are fractionating using salt slugs of 10ul each, with concentrations between 50 and 1000mM ammonium acetate.
The Switchos eluant is (2% acetonitrile/98% (0.05% TFA in water)).
The void peak has not shifted.
Thanks
The pH is 7.2 of the 1000mM ammonium acetate solution. We are fractionating using salt slugs of 10ul each, with concentrations between 50 and 1000mM ammonium acetate.
The Switchos eluant is (2% acetonitrile/98% (0.05% TFA in water)).
The void peak has not shifted.
Thanks
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Well, the pH is no problem. Since the void volume is unchanged, we can rule out de-wetting; not usually a problem with this system anyway. The 10µL injection volume is about equal to the void volume, so you may have retention loss and/or asymmetric peaks for the most hydrophilic peptides. Have you tested the column under standard 1-D isocratic conditions? If the column shows short retention times under standard conditions, then it is probably just worn out from loss of bonded phase.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
6 posts
Page 1 of 1
Who is online
In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 5 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 5 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science