Low flow-rates

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
So I've finally got gas running to my GC and she's conditioning my column right now. All my gas is running off bottles. The regulator for my carrier gas (N2) only goes up to 25 PSI. I'm only getting about 2-3ml min for my flowrate through the column. My column head pressure is 15 PSI. Is there an way to increase flow-rate, or do I just need a regulator that goes to 40?

Thanks.

Shorten your column.

Seriously, get a higher pressure regulator, unless you are only going to be using short columns. You need about a 10 psi delta between upstream and downstream for most regulators to control accurately.

Bruce Hamilton

Bruce is right.

A pressure regulator with a capacity of 60 or 100 psi is preferred.

Unless you intend to use short 0.25mm capillary, or 0.53mm ID, or 4mm ID packed columns ALL the time, use at least a 60 psi regulator.

best wishes,

Rod
3 posts Page 1 of 1

Who is online

In total there are 109 users online :: 0 registered, 0 hidden and 109 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 109 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry