Ran short sequence of clean samples in MeCl2. 1 uL splitless injections with 0.5 min 20 psi press pulse. Injector liner is clean Siltek-coated single gooseneck w/o packing. Injection speed 5 ul/sec. (My mistake!!)
Ended sequence in Stdby mode with oven at 200C, injector at 200C with 25:1 split and Xfer at 250C.
2 days later in Stdby mode, air-H2O check indicates system is dry, but filament emission check scanning m/z 30-150 shows MeCl2 spectrum. Ion time only ~2 msec & RICs ~16000.
How to proceed ridding system of MeCl2? I've not previously messed with carrier gas lines - and don't know if that's the right thing to do to start.
Thanks for suggestions.