Hi,

Our lab are using Empower 2 software. And we need help from anybody who has the hand-on experience of using this software to process the HPLC data to calculate the s/n value. We are not sure what's the right way to set the parameters to calculate the signal to noise ratio. We realized that different selection (base line, peak to peak, average peak to peak...) under the Suitability tab and the time window under the Noise and Drift tab end up with the so different s/n values. Help?
Peter

You need to make a custom field.

Signal to noise is peak height (in AU) / ave peak to peak noise in AU

The noise drift tab will calculate noise over the time you set in AU.

Peak height is AU times a multiplication factor,

So, your custom field should look something like this

(Peak Height / 100000)/Ave peak to peak noise

You will have to figure out the factor (I think its 100000 but I could be wrong)

Hope this helps.

AA
2 posts Page 1 of 1

Who is online

In total there are 551 users online :: 1 registered, 0 hidden and 550 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 550 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry