I haven't extracted the 537 compounds but I am finding that with the other methods like 544, you must do the step where you wash the bottle with solvent and use that solvent to elute the cartridge. Seems a lot of the analytes like to stick to the glass sample bottle and also in the transfer lines. I add the solvent, slowly rock the bottle on its side while I roll it so that the solvent touches every part of the glass, then let it settle to the bottom then use the same transfer line to pull it over into the cartridge to elute. I gained quite a bit of recovery with 544, 525.3 and 530 analytes this way. Makes for a longer extraction process but fixes the low recovery.Anyone have tips to share on the sample preparation for this method?Chromatography and MS wise, this method runs well. Nice peaks, good sensitivity, good calibration curves. We took precautions in terms of contamination and carry-over, no problems there so far.
However, in our first tests runs we have troubles with recovery (both spiked analytes and internal standards) for higher molecular weight compounds, especially the sulfonic acids. Recoveries are in the low % range for these compounds, while they're fine (60-100%) for the compounds that elute earlier.
Any particular tricks for the SPE protocol that we might be missing? Our calibration standards do not go through the procedure. With these recoveries, we're at the limits of sensitivity so we have to get it better before starting the validation.
Another possibility I have hear of especially with PAH compounds is to lower the gas flow in the evaporation step and increase the temperature a few degrees. Seems higher gas flows are worse in recovery than higher temperatures.Thanks James.
Our samples are in polypropylene recipients. I'm not sure if that makes them stick more or less compared to glass.
What you describe was actually tested yesterday. I was hoping that this would fix the problem but unfortunately there's no significant different in recovery...