I have to determine content of glyphosate in formulations (no postcolumn, no residua). Is possible to do it using UV detector? Can you recommend me a column for the determination?

Many thanks to all!

We are testing our glyphosate columns (PRP-X400) using an RI detector. Concentration of glyphosate and aminomethylphosphonic acid in our test samples is high. For UV detection prior derivatization might be required.
Probably some Pickering (or ex-Pickering, hello Mark!) can answer this question.
--
Robert Haefele
Glyphosate has no chromophore -> no UV-detection.

Commonly used is FMOC-derivatisation (precolumn), separation on standard C18 column and fluorescence detection. UV might work here if concs are high.

The post-column technique is nice for residue determination (EPA 547), but it is not really a good idea for formulation work. I would probably take Robert's advice.
Mark Tracy
Senior Chemist
Dionex Corp.

To Mark:

To take RI detection on Hamilton column? Is it possible to avoid RI detection and any other except UV? At the moment our budget restrict any purchase and we have UV and PDA detectors only.

For example, is it possible to use '- peaks'? To add a strong absorptive compound to the mobile phase and glyphosate will give negative peaks?!

To Robert:

Is it possible to determine glyphosate on both 79398 and 79473 columns?

Thanks to all!

There was an AOAC method for glyphosate using a UV detector in the 15th edition of AOAC methods. They used 195 nm. I think I have another published method using a UV detector but it is at home. Like you, we do not have an RI or post-column derivatization capability so we use a UV detector. We use 205 nm as it seemed less noisy (though the signal-to-noise determination was done by semi-calibrated eyeball). It is hardly ideal, but has been useable.

Blagoje

yes, both columns can be used for glyphosate analysis. The only difference is column id (2.1 and 4.1mm id, respectively)
--
Robert Haefele

I also used 205 nm for quantitation, during synthesis work. Whatman Partisil column with 20mM phosfate buffer.

Fabiano

We determined glyphosate with conductivity detection on the Metrosep A Supp 3 - 250 column (6.1005.320). Using a barbonate/bicarbonate eluent.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland

Glyphosate can also be determined using a Hypercarb (porous graphitic carbon) HPLC column from Thermo Scientific.
10 posts Page 1 of 1

Who is online

In total there are 20 users online :: 2 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google Feedfetcher and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry