hi
does pH of the sample injected matters(or affect) the stationary phase stability???

No single answer to that. If the mobile phase is well-buffered and the sample is small and/or dilute, there should be no problems. If your samples are large / high ionic strength / extreme pH, then a guard cartridge would be a good idea.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

hi
i have samples ranging from pH 3 to pH 9, the injection volume for these samples will be 100 microlitmy buffer is ACN+% TFA and water +0.1%TFA??
under these condition can the pH effect stationary phase

A guard is probably a good idea for this case, unless the stationary phase is rated for pH 9.
Mark Tracy
Senior Chemist
Dionex Corp.

My guess would be that your biggest problem could be pH incompatibility with resulting bad peak shapes (at the higher sample pH):
5 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Feedfetcher and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry