We are using an Agilent HPLC Series 1100 with an isocratic pump. We have a high pressure error with mobile phase only (THF). The pressure starts normally but increases in less than a minute. . The source of the high pressure seems to be in the pump area (column has been disconnected). We are using a brand new model Degassex4400 - with no pressure errors indicated.
We have inspected and cleaned the pump, replaced the active inlet valve cartridge, replaced the purge valve and frit, replaced prefilter frit, and replaced the capillary line.
The manual suggests the damper or LPM board. $$$$ ouch.
Any suggestions?
Thank you!