However, when I purge the machine as part of the start-up checks, the purged solvent mix comes out a yellowy/green colour. All the solvents are colourless when they enter the mixing chamber.
I initially had problems with the flow rate (set at 1.5ml/min but much less was coming out). However, after much purging this has resolved and the flow rate is now OK. The discolouration issue remains though and one of my baselines (when column is connected and purge valve closed of course) is very erratic.
Attempted solutions thus far have included purging at high flow rates (up to 5ml/min) with all possible ratios of mobile phase solvents, injecting 100ul of chloroform-enriched mobile phase in through the autosampler, vortex-mixing aliquots of the solvents together to check for colourimetric reactions. The problem occurs with all of the solvents, at 100% 0% 0% or at 34:33:33 or any other ratio. We were purging all afternoon with no improvement. Very boring and very frustrating.
I have a feeling that the discolouration and the erratic baseline are due to the same problem. Is is contamination inside the machine? The thing is there isn't really much between the mixing chamber and the purge valve so at which point can the MP turn yellow?
If anyone has any suggestions as to (a) what is causing the discolouration (b) what is causing the erratic baseline and (c) how to solve these problems, your advice would be much appreciated.
I'm a nutrition PhD student, not an analytical chemist (the in-house analytical chemist is equally stumped by the way), so please keep the jargon at novice level if possible.
Thanks very much,
Simon