Broad Peaks
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
21 posts
Page 2 of 2
What are your samples - we may be able to give you a clearer direction to head in about what sample solvent you may be able to use. How about trying 100% AcN or MeOH as the solvent?
-
- Posts: 10
- Joined: Wed Jul 11, 2007 2:03 am
Thanks Rob
My analytes are PAHs and I use isooctane to extract them from plant materials.
My analytes are PAHs and I use isooctane to extract them from plant materials.
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
I have tried to do that application. It is quite hard. Direct injection just won't cut it. The attempt to improve detection limits by large volume injection is defeated by the gross loss of efficiency.
If you are not afraid of complexity, there is an online SPE method that uses an exotic DACC concentrator column from Varian. This column will selectively extract PAHs from aliphatic solvents. After washing with iPA, it is switched inline with an analytical column (Varian again) for elution. Believe it or not, this is the basis of a well-established method in Europe. Dionex has an example implementation of the method: www1.dionex.com/en-us/webdocs/7503_summit_x2_poster_pah_application_isc04_sep042.pdf
If you are not afraid of complexity, there is an online SPE method that uses an exotic DACC concentrator column from Varian. This column will selectively extract PAHs from aliphatic solvents. After washing with iPA, it is switched inline with an analytical column (Varian again) for elution. Believe it or not, this is the basis of a well-established method in Europe. Dionex has an example implementation of the method: www1.dionex.com/en-us/webdocs/7503_summit_x2_poster_pah_application_isc04_sep042.pdf
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 374
- Joined: Fri Oct 22, 2004 7:14 pm
You can check out EPA method 8310 in which methylene chloride is used for extraction, then concentrated and finally exchanged to ACN.
Methylene chloride is a stronger solvent than isooctane, you can achieve better extraction efficiency by using MECL2.
By concentrating the extract, you can use smaller injetion volume for HPLC analysis.
Exchanging MECL2 to ACN will solve the immiscibility problem.
Methylene chloride is a stronger solvent than isooctane, you can achieve better extraction efficiency by using MECL2.
By concentrating the extract, you can use smaller injetion volume for HPLC analysis.
Exchanging MECL2 to ACN will solve the immiscibility problem.
-
- Posts: 1233
- Joined: Tue Oct 11, 2005 7:20 am
I'd agree with JI2002, unless there is a very good reason not to follow a standard procedure for PNAs ( eg EPA methods ), it's worthwhile to follow the method concept, even if you don't follow the detail ( such a the standard mixtures, internal stds etc ). Exchanging your solvent seems a very good idea.
As most PNAs are medium to low volatility, you should be able to cautiously concentrate your samples by evaporation.
The question about acceptable precision can only be resolved by the person using the the data. What's acceptable for a quick research scoping project probably will not be acceptable to a regulator being asked to use the data to grant consents or licences.
Please keep having fun,
Bruce Hamilton
As most PNAs are medium to low volatility, you should be able to cautiously concentrate your samples by evaporation.
The question about acceptable precision can only be resolved by the person using the the data. What's acceptable for a quick research scoping project probably will not be acceptable to a regulator being asked to use the data to grant consents or licences.
Please keep having fun,
Bruce Hamilton
-
- Posts: 10
- Joined: Wed Jul 11, 2007 2:03 am
Thanks a lot for all of you for the very nice discussion!
I think it is better to stop my game and follow one of the standard methods instead.
Thank you all again.
I think it is better to stop my game and follow one of the standard methods instead.
Thank you all again.
21 posts
Page 2 of 2
Who is online
In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science