I am analyzing by RP-HPLC a complex extract from plants. This extract contains a wide range of compounds, with huge variations in concentration and UV response and I need to quantify them. I have found out that, to achieve good peaks also for the minor components, I have to use a concentration of the extract which causes the major peaks reaching an absorbance of 2500 mAu…
I have read on the specification of the UV-detector (Agilent MWD 1100 series) that the linearity range is > 2 Au.
Can I trust it? Will my quantitative analysis be correct?

thanks
annavera

Even if it´s linear for your substances at such high absorption, you may have a strongly absorbing baseline such that you detector "sees black" at a highly absorbing sample. I don´t see how you can get around checking the linearity by dilution series.
2 posts Page 1 of 1

Who is online

In total there are 47 users online :: 3 registered, 0 hidden and 44 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 44 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry