the lot of the silica was different ?and the ACN used as mobile phase maybe from different company,I forget it!
testing conditions as follows:
mobile phase:50%ACN/50water
flowrate:50ml/min
wavelength:254nm
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
If you packed at 1000 psi this time and 750 psi last time, then all the conditions were not the same. And what does "about" 280g mean? Was that ± 1g? ±10g? There is an old saying that "the devil is in the details", and that certainly applies to chromatography columns.all the conditions were the same ,the same packing system,the same HPLC system,almost the same amount silica (C4,300A,10-15u,214TPB1015), about 280g,the same column hardware this time ,with higher packing pressure about 1000psi,last time the packing pressure is about 750psi.