We are seeing a codeine/syrup residue build up on the C8, 5 um, 150 x 4.6.

M.P. is 50/30/20 buffer/ACN/MeOH.

We are not analyzing for codeine.

Does anyone have a recommendation to wash the codeine and syrup build up from the column between runs?

Thanks :roll:

In the absence of more specific recommendations, try running a gradient from your existing mobile phase up to 60/40 ACN/MeOH and see what concentration of organic solvent it takes to elute your retained junk. Then periodically backwash your column with slightly higher organic.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

In addition to Tom's advice to get rid of codeine and other hydrophobic stuff, there may be gum thickeners in the syrup that are poorly soluble in your mobile phase. Gums are polysaccharides that are more soluble in water, so try a wash with 25:75 organic:buffer. If your column will tolerate reverse flow, do that.
Mark Tracy
Senior Chemist
Dionex Corp.

thank you both.
:P
4 posts Page 1 of 1

Who is online

In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry