- 
								
 
- Posts: 30
- Joined: Thu May 17, 2007 4:12 pm
Advertisement
Retention time(tR) and baseline peak bandwidth (Wb)
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
			2 posts
							Page 1 of 1
					
	
						Is there any mathematical correlation between the retention time of an analyte and the baseline peak bandwidth? I want to express the Wb of the peak as peak volume (Vc). I think that this can be calculated easily, but i would like some advice of how to estimate it.Is this equation correct ?           ( Vc=4Vm(1+k)/N^1/2  where Vm is the column dead volume, k is the retention factor and N the number of theoritical plates).
					
									- 
								
 
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
						Your equation is correct
					
									
			2 posts
							Page 1 of 1
					
	Who is online
				In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 32 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science














 
																							