Dear members of this forum,

My question is about the verification (operational qualification) of Liquid Chromatographer from Agilent Model 1100. In this step I used the new restriction capillary and the main difference with the old is the length, it is ten times higher. When I turned on the pump, the pressure increase a lot and I turned off inmediatelly. I want to know if anyone have experience in this step of verification (composition gradient) and until what maximum value of pressure can work without damaging the pump. This information will be very valuable for me.
Thanks in advance for your help,

Diego Delmonte
Dear Diego,
we use for OQ of our HPLC 1100 (Agilent) the same capillary, wich is used by Agilent: Part-No. 5022-2159. It works very well. Wich part no. do you youse?
The backpressure of the system can go up to 400 bar. This performs the leak test also.
I hope, this will help you!
Best regards!
Antje
2 posts Page 1 of 1

Who is online

In total there are 34 users online :: 2 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry