MG » Wed Dec 08, 2004 3:51 pm
If you look at the product ion spectrum of your ISTD (acetophenetidine), is one of the product ions the same m/z as paracetamol? Looking at the structures, it looks like this could be possible. If so, then paracetamol could also be formed as an in-source fragment of acetophenetidine, and then you'd see signal for the paracetamol transition at the retention time of the acetophenetidine. I've seen this sort of behavior before when working with structurally similar compounds. Since your analyte and ISTD are chromatographically separated, it is nothing to worry about, as long as your chromatography is stable. The only thing that worries me is your last statement:
The concentration of X doesn’t change if you increase the concentration of paracetamol or concentration of the internal standard
I assume you are referring to the response of X. I would expect the response of X to increase with increasing acetophenetidine, if indeed X is an in-source fragment of acetophenetidine.