esi/ms and polylactic acid

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I have a chemical engineer who is making polylactic acid (MW>100,000 Daltons). I understand the SEC instrument can be calibrated by getting the molecular weight of the synthetic polymer by esi/ms - preferably maldi-tof. Does anyone know if polylactic acid (MW > 100,000 Daltons) can be done by esi/ion-trap?
vestel b. shirley, president
betves inc
166 norwood drive
reidsville, nc 27320

Whether it will work, I can't say for sure, but it is possible. To see it by ESI-ion trap, it will have to form multiply charged ions within the m/z range of your trap. The upper limit of your trap is probably 4000 to 6000 m/z. Also, I don't know what that polymer is soluble in, but if it will only dissolve in a nonpolar solvent, then you'll have a better chance of getting signal if you add 5-10% of a polar solvent, such as isopropanol. Sounds like an interesting experiment, and I would like to hear how it turns out if you attempt it!
2 posts Page 1 of 1

Who is online

In total there are 40 users online :: 1 registered, 0 hidden and 39 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 39 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry