Ip address Agilent GC 7890a MS 5975c
2 posts Page 1 of 1
Hi all,
we have an Agilent GC 7890A / MSD 5975C connect to a PC (HP Agilent) but we need to change the Ip address . How can I do that? or where can I find the instructions to do that?
Thanks.
Hi all,
Thanks.
In fact there are 3 different IP addresses.
GC itself, MSD and computer with ChemStation (or whatever CDS you have).
You may change GC IP from GC keyboard.
You may change MSD IP address from it's local control panel.
And last, PC IP is changed during instrument configuration.
All 3 IP addresses must be different.