Can anyone suggest a solution?
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Thanks for the quick response. May be TGA-MS can serve the purpose. Only thing is if 2-3 gases evolve at the same temperature, then may be it is difficult to find the exact amount of each gas evolved. That is why, we thought about TGA-GC-MS.I'm not sure what you mean by TGA-GC-MS can't go that high. When you are above 400C you will be pyrolyzing compounds, and the molecular weight of most of the fragments will be low enough to pass through a GC column into a mass spec. Of course the transfer line would have to be hot enough to prevent condensation, but the same would apply to TGA-MS.
Ron:As long as you have masses for each gas evolved that are not in the other evolved gases you should not need separation to quantitate the gases