I learned that nonporous reversed-phase column with smaller particles(1.5um) have some advantages,such as higher recovery for protein analysis.We have a Mariner ESI-TOF and a common HP1100 HPLC systerm, I don't know if this column can work well in our instrument. I know that TFA may cause some supression,but I don't know if it can destory the instrument equiped needle tip in our ionspray sourse.I would appreciate it very much if someone could give me some advise.
-------------------------------------------------------------------------------------------------------
By Anonymous on Friday, December 26, 2003 - 03:14 am:
In principle no need to use TFA. Use formic, is as good. Works with porous, but do not know on nonporous.
-------------------------------------------------------------------------------------------------------
By dingyuan on Sunday, December 28, 2003 - 08:29 pm:
Thank you for your concern.I think porous is avaliable for peptide analysis rather than protein.Would you please give me more concrete information.
-------------------------------------------------------------------------------------------------------
By MG on Monday, December 29, 2003 - 06:48 am:
I've used TFA in mobile phases periodically at concentrations less than or equal to 0.2% (v/v) with LC/MS. To my knowledge, it never damaged my ion source. However, it will cause ion suppression, and will stay around as background even when you think you have flushed your system adequately.
-------------------------------------------------------------------------------------------------------
By dingyuan on Tuesday, December 30, 2003 - 12:33 am:
Thanks.We have destroy our tip using chloroform as solvent and think halogen may have bad effect.
That prevent me from drawing a rapid decision.
-------------------------------------------------------------------------------------------------------
By MG on Tuesday, December 30, 2003 - 06:06 am:
Is your ESI needle stainless steel? I would be surpised if chloroform caused damage, although you might not get much ionization.
-------------------------------------------------------------------------------------------------------
By dingyuan on Thursday, January 1, 2004 - 02:39 am:
Yeah, a stainless steel tip is destroyed when using chloroform as solvent. I think halogen may cause some problem.
-------------------------------------------------------------------------------------------------------
By Uwe Neue on Thursday, January 1, 2004 - 07:50 pm:
Symmetry C18 columns work very well with peptides and proteins with formic acid as the modifier. They are available in a 10 nm variety for peptides and a 30 nm variety for proteins. Recovery is without difficulty.
-------------------------------------------------------------------------------------------------------
By dingyuan on Friday, January 2, 2004 - 07:06 pm:
Is sysmmetry C18 a waters product? I obtained some information about Vydac, Jupiter, BioBasic, Discovery BIO Wide Pore.I hope someone could kindly give me some advise.
-------------------------------------------------------------------------------------------------------
By Basil on Wednesday, January 7, 2004 - 04:26 am:
Yes dingyuan, Symmetry is a Waters product..
I've been working with a Symmetry C8 for a developing new methods and now I'm going to buy a C-18 for an specific one.
Hope it helps
Basil
-------------------------------------------------------------------------------------------------------
By dingyuan on Saturday, January 10, 2004 - 01:09 am:
I went away some days. I am pleased to recieved your recomments.Thanks.
-------------------------------------------------------------------------------------------------------
By Anonymous on Wednesday, June 9, 2004 - 11:02 am: