Clearly adding a guard column as you describe would give an improved plate count initially simply by having a longer column. I am sure the suppliers would be very happy to supply guards on such a premise but I would not be an advocate of this. You assume that the negative effects associated with column coupling and increased analysis times through use of an extended bed length are insignificant. Of course with dirty samples the increased plate count would be lost if the guard column rapidly becomes contaminated.
The reality of every day operations means that guard columns are used for convenience rather than as an essential element. The use of guard columns may well be unavoidable for the type of samples you handle. The point I make is that the use of guard columns should be regarded as a method of last resort and in the situations you describe they are the lesser of 2 evils