Dwell volume with MS detector?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
This might be a dumb question :oops: , but all the methods I know for measuring dwell volume involve a UV detector, running a gradient from 0-100% A to B with B having some UV active component like acetone. I have a mass spec. How can I measure dwell volume with a mass spec?

The easiest thing would be to borrow a UV detector for the morning and use that (all the dwell volume is pre-detector anyway).

If you really want to use the MS, you can adapt the UV procedure. Find a compound that that you can detect nicely and for which you have a broad linear range.

In your A reservoir, put a solution of that compound at a low level (just above the LLOQ). In the B reservoir put a solution of the same compound at a higher level (near the upper end of your linear range). Run a gradient from 0 - 100%B and monitor the signal. The computation is identical to that for the UV detector.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 38 users online :: 0 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry