Residual Sum of Squares

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

9 posts Page 1 of 1
Within linearity validation, ICH guidelines ask for reporting of the “residual sum of squaresâ€

It provides a measure of the scatter in the data. I'm not enough of a statistician to give an opinion as to whether it's better/worse/equivalent to r^2 in that regard.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
It gives an idea about how close the experimental values are to the true value.

I'm not an expert on statistics, nor familiar with ICH protocols, but to elaborate a bit (and share my bits of info from a recent regression study) ; residual sum of squares is also known as "unexplained sum of squares", or "sum of squared errors". Well known Pearson's r squared is calculated as 1-("sum of squared errors"/"sum of squares").

Sum of squares is easily calculated by adding up squared "each y value minus mean of y values".

From my understanding residual sum of squares is no better than Pearson r as being a measure of scatter.
I can also repeat beforesaid warnings: one of the main assumptions of least square is that errors are "homoscedastic". That means reasonably similar along the range, which is not the case for chromatography. But if we study a narrow range, this assumption may be validated.

Hope, didn't make things more confusing.

Best luck,
Bulent

Thanks for your answers so far.

What I was referring to with my expression “absolute valueâ€

The sum of residual squares gives a better measure for the quality of a calibration curve than r^2. Especially in cases where the calibration curve is not strictly linear.
If the slope of the calibration curve is continuously increasing or decreasing with increasing concentration, you may get a perfect r^2=0.9999 but the sum of the residual squares is quite large. In such a calibration the calibration points are not statistically spread around the curve. But the highest and lowest points are below and the points in between are above the curve (or vice versa).
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland

The difficulty lies in the interpretation of sum-squared-residuals. What do you compare it against? How do you know if it is "too large?" SSR will increase with the number of data points. It depends on the units of Y (try converting from mAU to AU and see what happens). The correlation R^2 is an attempt to normalize the SSR by dividing it by the degrees of freedom (n-2) and the variance of Y. This removes the two effects and allows comparisons to be made between different calibrations. Unfortunately, neither SSR nor R^2 is a satisfactory measure of linearity.
Mark Tracy
Senior Chemist
Dionex Corp.

> Unfortunately, neither SSR nor R^2 is a satisfactory measure of
> linearity.

As a measure of linearity, we use the relative standard deviation of the response factor (that is peak area divided by concentration).

Do you think this is a good measure or is there a better parameter?

Hartmut

The RSD of the response factor suffers from the same defect that R^2 does: it treats non-linearity and noise as the same. In other words it can tell you that the line is a poor fit to the data, but it can't say what is wrong. The poor fit can be due to either the data being too scattered or the calibration function is the wrong choice (or both).

At the moment, a sharp eye and a trained mind are the best way to recognize nonlinearity.
Mark Tracy
Senior Chemist
Dionex Corp.
9 posts Page 1 of 1

Who is online

In total there are 37 users online :: 1 registered, 0 hidden and 36 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 36 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry