hi

I´m an undergrad student doing a practical internship
5 weeks ago I had never ever dealt with a GC, not to speak about fixing problems...

when I can see clearly the peaks, but the integrator does not integrate it, is one possibility (among others!)of the problem a wrongly set Peak Witdh?
does changing the peak witdh influence the calculating? I mean do I need to make a new calibration after I changed this parameter?

when I plot the sensitivity against the detector current, does the graph pass through a maximum or is it all the way increasing until the filament burns out?

and I think my GC (Gow Mac 550 Series) has some kind of slacky joint or the like... without any warning the baseline becomes a straight line without any noise
running with sample gases I only get idle peaks
the peaks are in the right place, but sometimes they even go in the wrong direction
when I switch the polarity then the plot looks like normal, just upsidedown of course
do you have any idea what it could be

I´m lost...

thx
Michael

you probably need some more info

it deals about detecting H2, CO and CO2 in He carrier gas flow
I use a TCD detector and a HP 3396 integrator

H2 detection is tricky, range 0...20%

Michael,

I'm sorry I am not familiar with your instrument, but in general, peak width, slope sensitivity, and area reject are all things that may need adjusted in your integration parameters. Also a calibrated peak may not be identified properly if it is outside the retention time window of your standards. Ideally when you correct your integration problems you should re-calibrate, or re-integrate stored calibration runs with your new integration parameters.
mischka80

You should not try to measure Hydrogen in Helium at levels 5-10% The response of hydrogen changes and it behaves non-linearly.

Use Nitrogen (or Argon) carrier to measure hydrogen.

best wishes,

Rodney George
rgeorge@sial.com
mischka80

Sorry I forgot to add it appears that your TCD is not balanced. You have the flow through the reference at a level too high or low compared to the measurement cell. Try to find the manual for the detector and look at the examples for troubleshooting.

:)
5 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry