As you can pump and inject almost everything what is liquid with conventional HPLC, it is a different story with a tuned UPLC system.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Did you bother reading the white paper I referred to?. Nowhere did I, or the paper, suggest the 2.5 um column would exceed the theoretical plates of a smaller diameter particle. What I quoted was that the theroretical performance gain of sub-2um particles, calculated using particle size, was not achieved.Sorry to be rude, but I have the feeling that things are mixed up here.
Like Uwe mentioned, it's never possible to get the same results on a 2.5um column as on a 1.7um column.
Do you mean radial temperature gradient Uwe?In addition, there are a lot of additional complications in measuring platecounts on very tiny particles. In principle, plate count measurements become meaningless, if the pressure is high enough