As you can pump and inject almost everything what is liquid with conventional HPLC, it is a different story with a tuned UPLC system.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
 
   Overpressure. One thing we have noticed is that if we lower, to say 0.05ml/min to conserve mobile phase when not running or stop the flow, the pressure build up on the next injection is immediate. One idea we are therefore trialing is to keep flow constantly going through the system in an effort to minimise any pressure build up - Will keep you all posted!
 Overpressure. One thing we have noticed is that if we lower, to say 0.05ml/min to conserve mobile phase when not running or stop the flow, the pressure build up on the next injection is immediate. One idea we are therefore trialing is to keep flow constantly going through the system in an effort to minimise any pressure build up - Will keep you all posted!
					 We looked into the columns that developer1974 recomended earlier in this thread. We decided to trial the Phenomenex Luna HST C18(2) 2.5µm 50x2.0mm as a substitute for our poor performing Waters Acquity BEH C18 1.7µm column.
  We looked into the columns that developer1974 recomended earlier in this thread. We decided to trial the Phenomenex Luna HST C18(2) 2.5µm 50x2.0mm as a substitute for our poor performing Waters Acquity BEH C18 1.7µm column.  The chromatography was near identical to what we had achieved with the Waters column. Backpressure was down to around 4,000psi pmax and we have sucessfully run >1000 injections with no pressure increases whatsoever.
 The chromatography was near identical to what we had achieved with the Waters column. Backpressure was down to around 4,000psi pmax and we have sucessfully run >1000 injections with no pressure increases whatsoever. 
					 
  
 Did you bother reading the white paper I referred to?. Nowhere did I, or the paper, suggest the 2.5 um column would exceed the theoretical plates of a smaller diameter particle. What I quoted was that the theroretical performance gain of sub-2um particles, calculated using particle size, was not achieved.Sorry to be rude, but I have the feeling that things are mixed up here.
Like Uwe mentioned, it's never possible to get the same results on a 2.5um column as on a 1.7um column.
Do you mean radial temperature gradient Uwe?In addition, there are a lot of additional complications in measuring platecounts on very tiny particles. In principle, plate count measurements become meaningless, if the pressure is high enough