Headspace Pressure Issues

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi there,

I'm trying to run GC-HS analysis and I'm having pressure issues on the HS side. The GC (Clarus 500) is fine but the HS (Turbo Matrix HS40) pressure is higher than the one set.

The set pressure is 15psi but it would not go lower than 19psi and this after waiting ages. I checked for leaks at the back of the instrument & gas connections but this seems to be fine. I was wondering whether someone had any idea what the problem could be & if possible how it could be fixed?

The GC carrier flow rate is 5 ml/min with a split flow of 10 ml/min & the carrier is Helium.

Thanks.
jane

What pressure is on GC? I f you have there 30 psi, then you will not have 15 psi on headspace ever...... :wink:

Hi Jane

Depending on your column diameter, and whether you are setting pressures or flow rates on the GC and/or headspacer you will be forcing the pressure up by trying to squeeze too much gas down the column.

Peter
Peter Apps
3 posts Page 1 of 1

Who is online

In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry