Tom,
buffer concentration is 0.01 M phosphate buffer pH 3.85
samples are dissolved in 0.05M phosphate buffer:ACN (90%:10%)
concentration of the samples is .0048mg/ml
inj voume 50uL
only one degradant peak is tailing
There is a parent compd and 5 degrardants.
the degradants are
cyclohexane diacetic acid monoamide
Carboxymethyl-cyclohexanecarboxylic acid
3-Azaspiro{5,5}undecan-2,4-dione
2-Azaspiro(4,5)decan-3-one
1,1 cyclohexanediacetic acid
{1-(3-oxo-2-aza-spiro[4,5]dec-2-ylmethyl)-cyclohexyl]-acetic acid.
polarities are range, the last one being the least polar.
Please respond.
Thanks in advance.
First of all, phosphate is not an effective buffer at pH 3.8
Second, tailing of the parent is fairly common when looking at degradants, because you have to overload the parent compound in order to detect the degradants.
Beyond that, you really haven't provided enough information:
what is the buffer concentration?
What is your sample dissolved in?
what concentration is your sample?
what volume are you injecting?
are any of your analytes (the parent or degradants) ionizable? if yes, acid or base? can you estimate the pKa?
Do all the peaks tail, or only the parent compound? And when you change columns, do you see the tailing/splitting immediately or only after some time?