I once encountered distorted peaks in the analysis of sterol TMS derivatives using hydrogen as a carrier gas. There were never any problems when using helium. I thought this might be due to some sort of hydrogenation of the sterol double bonds at the high temperatures of the system (around 250 degrees). If this were true, it might indicate potential problems at the column end as well as the mass spec end. However, it could be a rare example but was definitely attributable to the use of hydrogen.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Hi Victor
Was it the chromatographic peaks that were distorted, or the mass peaks ? If the sterols were hydrogenating I would think that there must have been some dirt in the system that was acting as a catalyst. Come to think of it I vaguely remember seeing something about hydrogen reducing metallic oxide dirt in hot inlets.
Peter
Was it the chromatographic peaks that were distorted, or the mass peaks ? If the sterols were hydrogenating I would think that there must have been some dirt in the system that was acting as a catalyst. Come to think of it I vaguely remember seeing something about hydrogen reducing metallic oxide dirt in hot inlets.
Peter
Peter Apps
-
- Posts: 308
- Joined: Tue Aug 31, 2004 12:26 pm
Hi Peter,
This was definitely a chromatography problem as I noted it on a GC-FID system. I think you could be right about dirt catalysing such a reaction- or it could be residues of materials used in e.g. column deactivating processes (during column preparation) that could provide catalytic sites.
This was definitely a chromatography problem as I noted it on a GC-FID system. I think you could be right about dirt catalysing such a reaction- or it could be residues of materials used in e.g. column deactivating processes (during column preparation) that could provide catalytic sites.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Metal components used in MS detectors in the presence of hydrogen will hydrogenate double bonds.
Hydrogen carrier gas does not exactly reproduce the spectra of analytes recorded in databases and used for many legal issues.
Legal issues are the primary reason hydrogen carrier is not used for MS spectra.
However, for pure research and cost savings I do not see a problem.
Just don't submit any papers to journals using MS with hydrogen carrier.
At least, that is my advice.
best wishes,
Rod
Hydrogen carrier gas does not exactly reproduce the spectra of analytes recorded in databases and used for many legal issues.
Legal issues are the primary reason hydrogen carrier is not used for MS spectra.
However, for pure research and cost savings I do not see a problem.
Just don't submit any papers to journals using MS with hydrogen carrier.
At least, that is my advice.
best wishes,
Rod
19 posts
Page 2 of 2
Who is online
In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 11 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 11 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science