Dose anyone know the brand of HPLC column which can bear high concentration of salt without reducing of lifetime.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
HIC (hydrophobic interaction chromatography) columns, for example, are designed to be used with 4N sodium sulfate, and most ion exchange columns can used with >1N sodium chloride.
Yes, I'm being sarcastic.
That question is unanswerable without more information like:
- mode of chromatography ?
- intended pH ?
- what salt ?
- organic solvent ? which one? how much?
- are you referring to salt in the mobile phase or in the sample?
Yes, I'm being sarcastic.
That question is unanswerable without more information like:
- mode of chromatography ?
- intended pH ?
- what salt ?
- organic solvent ? which one? how much?
- are you referring to salt in the mobile phase or in the sample?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Let us assume that you are talking about reversed-phase chromatography: for these columns, a high salt concentration in the mobile phase does not affect lifetime (assuming pH is within the allowed range for the column). If you are talking about injecting a high salt concetration on such columns, the problem is probably a different one: precipitation of salt. No column likes such a thing...
Tom is right: your question needs to be clearer...
Tom is right: your question needs to be clearer...
-
- Posts: 33
- Joined: Sat Jan 20, 2007 10:44 am
I think Neue and tom is right.
As you can use high salt in revrse phase only in specified pH and morever you can give washing to column extensively as you will finish your work. If you will give washing first 100% water followed by solvent ratio in which your generally storing column.
As you can use high salt in revrse phase only in specified pH and morever you can give washing to column extensively as you will finish your work. If you will give washing first 100% water followed by solvent ratio in which your generally storing column.
4 posts
Page 1 of 1
Who is online
In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 24 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science