The samples I am injecting is protein which is acid hydrolyzed by 4.5M TFA to monosaccharides. The hydrosylates are analyzed using a Dionex ion exchange chromatography system with a Pulsed Amperometric Detector (DX500 model). The peaks of interest are glucosamine and mannose. A Dionex CarboPac PA20 column and same guard column are used, and the peaks are eluted isocratically with 10mM NaOH.
My concern right now is that I don't seem to be able to quantify correctly any sample that started with protein in it. I have tried injecting protein sample with a spiked amount, along with a protein sample without the spiked about, and also a placebo with the spiked amount (n=5). The placebo with the spiked amount, which contains no protein, quantitates correctly as expected (1nmol). But, when I add the nmol amounts for the the placebo spiked (1nmol) and the protein sample not spiked (1.22nmol), the amounts do not equal the protein spiked sample amount (1.76nmol). When I do this on two instruments, I get similar results, where the placebo spiked is the same nmol amount, but the protein samples are about 40% different between each instrument.
Does anyone have any insight as to whether the protein would have any effect on the column over time, or if the TFA which is evaporated and reconstituted with water would have any effect? Would the protein have any effect in the detector cell if one were cleaner than the other? I know the column and instrument are good for analyzing sugars, but do they work fine when injecting protein along with the sugars? Please let me know if anything jumps out as the culprit. Thanks!