Autosampler thermostat verification

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
We have recently purchased Agilent 1200 thermostatted autosamplers and are updating out SOP to include a temp verification. We were going to measure the thermosat temp at 4 and 20 oC at 5 points in the tray. Does anyone else perform a similar verification?

Dear mojo:
We never do any temp verification for the autosampler. If the instrument is located in a temp-controlled environment, it is not needed.
Alfred.

We used a small temperature probe that fitted inside uncapped vials filled with water and tested at our standard temperature, 4, 10, 20C. I think we only tested one corner and the centre, as it takes a while to reach thermal equilibrium after opening the door and moving the probe.

As most of our methods were developed before we had thermostated autosamplers, we also set quite large acceptance tolerances for any methods that specify temperatures other than ambient. That's because if we filled the tray with vials, it takes a while for the contents to stabilise, unless the vials were precooled in a fridge.

Bruce Hamilton
3 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry