the method for the analysis of alendronate by FMOC derivatization in pH 9.0 borate buffer and then extracting in DCM works perfectly but the area of the alendronate peak varies evertime. is this method reproducible , according to me it is difficult to get same results
shweta chauhan
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
When you say the area varies, do you mean for the same vial you get different areas depending on the time of injection? Is the variation random or is there a trend?
I have found that to get reliable FMOC derivatization, you need rapid and vigorous mixing of the reagent with the sample.
I have found that to get reliable FMOC derivatization, you need rapid and vigorous mixing of the reagent with the sample.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
2 posts
Page 1 of 1
Who is online
In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 33 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 33 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science