SIL-10AXL injecting air

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

21 posts Page 2 of 2

DR,

I will replace these parts when I get them in. There was no ball. The front plate comes off with the stator in it, then the rotor connects via two pins to a stainless steel surface that does the turning. I suspect the bearing (ball) is behind this. Thanks again to everyone who contributed! I will let you know how it all comes out.

John

Still no luck. The air injection is still randomly occuring. I have replaced every consumable the company reps and I can come up with. I am going to move on to another piece of equipment for the time being. However, if there are any other suggestions I would be happy to hear them. Thanks again for all teh help!

No sure what the differences are between the 10A and 10AXL, but we had sporadic air problems with a 10A autosampler. After much aggravation we set the system up to make multiple injections from a vial filled with solvent and looked for air bubbles in the lines. Eventually we saw them in the line that appeared to have been "bent". While there was no leakage from this line, we replaced it and our problems stopped. Since nothing else seems to be working, I thought I'd throw this out.

Russ,

It sounds like we are having the same problem. How did you know which line the air bubble was coming from (my system is plumbed with all stainless steel)? Also, by "bent", do you mean that the tubing had just been bent by accident? Finally, which piece of tubing was your problem? I have replaced the tubing connecting the three port to the six port valve and the new tubing seemed to acctually make it worse! Thanks for your help.

John

John

Our system uses "plastic" lines from the solvent reservoir to the syringe unit and from the syringe unit to the 6-port valve assembly. I think our problem was with the line from the solvent reservoir to the 3-port valve of the syringe unit. We would sporadically get air into this line which would eventually make its way to the injection valve. While watching the system make multiple injections (a fun job) we noticed air getting into this line and it appeared to be coming from the area where the line had been accidentally bent. Replacing this line fixed the problem, though at the time I did not have a lot of hope that it would. However, these lines were plastic on our system and not stainless steel, so we were able to see the bubbles. As I remember it, we spent a lot of time looking at the 6-port valve and the 3-port valve of the syringe unit when our problem was "upstream" from these valves. Our system has an SIL-10A autosampler but I saw a picture of an autosampler listed as an SIL-10AXL and it appears to be similar. Good luck.

Russ,

The inlet to the syringe unit is plastic on my system, but tubing from the six port to the syringe unit is stainless. I have replaced this piece of tubing, but not the inlet. I will give the inlet tubing a try. Thanks for your input,

John
21 posts Page 2 of 2

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry