We recently bought a new Thermo Surveyor HPLC system to replace an aging Thermo TSP system. However, transferring a method from the TSP system to the new Surveyor system resulted in the appearance of a new system peak that coelutes with a key analyte. The mobile phases contain acetonitrile, water, acetic acid, and triethylamine at pH 10; separation is by C18 and detection is by fluorescence.
By substituting various Surveyor system components with the TSP counterparts in a process of elimination, the source of the contaminating peak was narrowed down to between the solvent proportioning valve and the pumps, inclusive. The mobile phase, solvent tubing, degasser, injection valve, tubing from the sample vial to the injector valve, and column were all eliminated as possible contamination sources. As the contaminant elutes as a distinct peak, it must be entering the flowpath upstream of the column exit and being resolved on column; therefore plumbing downstream of the column and the detector were not considered as contamination sources.
This leaves the proportioning valve and pump chambers as the only possible sources of contamination. The proportioning valve materials, check valves, and pump seals are made of the same material as on the TSP system. However, in the Surveyor, the pistons are made of zirconium ceramic; someone suggested to me these might be the cause of the problem - the pistons in the old TSP system are sapphire. I can get replacement sapphire pistons for the Surveyor, but are zirconium pistons known to be problematic for some analyses??
thanks
Tony