We used the Waters Oasis 2x4 strategy and found following method:
Oasis MCX cartridge,
condition/equilibrate with 1 ml MeOH, followed by 1ml water.
Then 1 ml sample load. Then there is a wash with 2% formic acid. And a elution step with 100% MeOH. After this, amoxicillin was eluted with 5% ammonia in MeOH.
Our samples were clean, no matrix effects at all, but: after 30 minutes, our amoxicillin was gone. This was due to the high pH of the ammonia solution. We then switched to 1% ammonia in MeOH, and before eluting the samples, there was enough 2% formic acid in the tubes to lower the pH from 11 to 4.5.
Unfortunately this was'nt enough and we still get recoveries of 66% instead of 100%.
We also tried HLB cartridges (reversed phase), but the amoxicillin elutes in the wash step.
Anyone who has some ideas?
Thanks