Thought I'd share what I think is a useful lesson that I learned this week while quantifying caffeine using a photodiode array and ELSD in series.
The samples: 3 Prototype creams with significant concentrations of caffeine to be quantified.
The goal: Quantify the caffeine as requested and provide a fingerprint of each sample identifying as many of the other major components as possible (excipients, preservatives, other actives, etc...).
The result: Calibrations for both detectors (6 points from 0.25-2X dose) RSQ > 0.999. ELSD calibration was quadratic, of course. System suitability (area) for PDA = 0.34%; for ELSD = 1.36% (N=6).
PDA: Sample recoveries were perfectly on target: 99.5-101% recovery on all 3.
ELSD: Sample recoveries ranged from 80.3-81.1%
What HAPPENED??
After a few moments of head scratching, it came to me:
1. The caffeine standards were dissolved in MeOH-H2O (1:1) whereas the creams were dissolved in neat MeOH in order to pull as much of the rest of the sample (preservatives, other actives, lipids, etc...) into solution and to precipitate any water soluble thickeners (carbomer, etc) that may be present.
2. The caffeine peak is the earliest eluter in the method.
3. My system volume is relatively small compared to my injection volume.
4. The ELSD is a non-linear detector.
What happened, I believe, was that there was slight band broadening with my samples that was not present with the standards. This was tough to see at a quick glance because my sample runs were 30 minute gradients whereas my standard runs were 5 minute isocratic runs. The sample gradient has a 5.5 minute isocratic delay at the start in which the caffeine elutes, so I can get away with that. Close inspection of the sample chromatograms revealed a slightly rounded peak for caffeine! That band broadening didn't effect the PDA results to any great degree, but because the ELSD is non linear, it really did a job on those results. Peak purity, by the way, was fine and consistent between the standards and the samples.
Preparing some standards in neat methanol and recalibrating with slightly broader peaks on the ELSD confirmed the theory - recoveries mirrored the PDA results perfectly.
Dropping the injection volume from 10ul to 5ul for samples and standards also brought peak shapes more in line.
I'm considering incorporating SPE into this particular sample prep should these prototypes move much further forward.
Lesson learned! Hope you don't mind my sharing my goof du-jour!