ATD Turbomatrix 150: Cleaning the Transfer line

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello all,

I have been using the ATD150 a lot recently and have been getting back a lot of impurities and unknown peaks. I recently found my trap had become very dirty and replaced it, expecting that to clean up the chromatogram, but the problem persists. I want to check if there is more contamination in the transfer line, but am unsure how to best go about cleaning it out.

Does anyone have any suggestions on how to clean the transfer line?
Buy:
1) deactivated fused silica column thats the same id as your gc column
2) ferrule for fitting in turbomatrix (see manual)
3) column union

Bonus: some 1/16" ptfe or pfa tubing to feed through the PE transfer lone, this will help protect you column when you feed it through. Ptfe fittings start to flow at 260C, do not go above 275 in your transfer line with the ptfe.

If you trap is visibly dirty, replace the transfer line. Lost thermally labile peaks, replace the transfer line.
@tdgcuser

Thanks for the suggestions, I will look into these!
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry