This is probably basic question for most of you (apologies for that!), but I would like to know how to deal with the trendline from my linearity test going through the y-intercept.

i.e. To determine the response factor for my compound by HPLC, I measure the absorbance of a range of concentrations then construct a graph which gives a straight line. The line has a good R-squared value and I use the gradient as my reponse factor for measuring samples. However, the line passes through the y-axis above the origin. How should I deal with this in my calculations.

thanks,

Nick.

scroll down this forum to page 2 and you'll find a thread called "Importance of Y-intercept in Method Validation"

http://www.sepsci.com/chromforum/viewtopic.php?t=4906

First of all, how non-zero is the intercept? Your data package should calculate a standard error for the y-intercept. If zero is included in that range, then the intercept is statistically indistinguishable from zero (in other words, don't worry about it).

If the intercept is significantly above zero, what you do depends on how you are using the results.

The most stringently regulated case applies to pharmaceuticals; even there, if you are doing something like drug purity or content uniformity, you only have to demonstrate linearity in a range bracketing your actual concentration (in other words, don't worry about it). If you are doing trace analysis, then a significantly non-zero y-intercept is something to worry about.

In no case should you "force" the intercept to zero.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry