- Posts: 535
- Joined: Fri Jun 02, 2017 11:10 pm
Hmmm, I found the location where pentachloroethane should elute and there is not even a trace in any of the samples. Is it possible it completely converted? I am going to contact the PE provider and ask them if pentachloroethane was present.It will usually be Penatchloroethane that decomposes into Tetrachloroethene, and Pentachloroethan does break down quite easily at room temperature.
If you only have a few samples to run at a time you can store them in the cooler at 4c and that will retard the breakdown for a few hours while it warms to room temp on the autosampler.
I still have the original stock. I'll prep it again and run it just as the autosampler is grabbing the vial.