I am very new at GC but the facility I am managing does have an analytical side, which I am also in charge of. I've been trying to optimize a test separation and I've read an awful lot online and in a few books but I needed some clarifications on a few points...
(1) I am using an Agilent 6850 with FID. I was injecting a compound with a bp of 265 C, dissolved in MeOH (I'm aware I didn't pick the right solvent). I set the detector and inlet T at 285 and the oven at 60 (isothermal). I'm using a 30m, 0.32 column with helium as my carrier gas. I set the inlet pressure to 8 psi (according to a table), split injection of 500:1 (fairly concentrated sample) and 1 uL injection.
According to Restek's backflash calculator, this would lead to excessive backflash. I have read in some places that backflash does not apply to split injections but I also have read that it does... Should I be concerned?
(2) I found that setting a higher pressure (25 psi, like was suggested bv previous users) led to peak tailing (although I had to use a lower split ratio of 100:1 due to the higher pressure). How should one go about adjusting pressure and flow rate for optimization?
(3) Pulse pressure related to injections - contains the sample expansion and transfers the analytes to the column faster - is this something I should consider doing, especially when using MeOH as solvent?
(4) What three top solvents are generally the best to use for dissolving samples (once solubility is assured)? Do you use a high solvent grade for GC-FID?
Thank you so much for your help,
Roxanne.