Autosampler wash needle depth question 7683/6890N

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
So, my problem is that the needle on the wash only goes down half way into the wash vials. It goes about to the same depth as the sample vial (which is raised up compared to the wash vials--as seen in picture). Because the wash doesn't use all of what's in the vial, we can't do long runs without going in and refilling them during the run.

Is there any way to set the needle to penetrate farther down into the wash vial? I couldn't find anything in the manual, except for 'sample offset', which I believe changes how high above the sample vial the needle goes-this sounds good but I think it is only for the sample vial itself and not the wash vial.

If this is just the way it is, I can accept it and maybe we can just wash less, but it seems to me they wouldn't design it to work like this. This is on a 7683 HP injector and 6890N Agilent GC.

Image
It's by design. If you have original wash vial you can see a level marked on the wall. This is the minimum solvent level which is ca. 18 mm from the bottom.
Image
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry