why the detection wavelength changes ....?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Dear All,
Why does the detection wavelength changes for the same drug product for a particular analysis( say, dissolution for instatnce) when analysed via HPLC euipped with UV detector and secondly analysed on UV-Visible Spectrophotometer.
Thanx in advance.
Shrikant/Harsha
Dear All,
Why does the detection wavelength changes for the same drug product for a particular analysis( say, dissolution for instatnce) when analysed via HPLC euipped with UV detector and secondly analysed on UV-Visible Spectrophotometer.
Thanx in advance.

Ideally it should not change if your diluent (blank) for the drug is kept same for both HPLC-UV and UV analysis.
lambda max varies if you are changing the diluent.
i.e. if you are taking acidic diluent you will get some lambda max but for the same drug lambda max will vary if you chose the basic diluent

Its not necessarily b/c lambda max is changing. The HPLC method may have been optimized by changing the wavelength. It may be that lambda max caused high baseline noise. Maybe the wavelength selected in the HPLC analysis was not the lambda max for the active but useful for evaluating other related compounds or excipients allowing additional evaluation in the analysis without sacrificing the detection sensitivity of the active.

the dude
"Take it easy man....there's a beverage involved here"
Harsha,

The differences you are seeing seem to be related to two instruments (Uv detector, Uv-Vis Spectrophotometer). Therefore thevwavelenght accuracy is likely to be different in each. Obviously, large differences, such as 10-20 nm, are most likely due to solvent effects, temperature, etc.

I hope these comments help you.

josebenjamin

Method Optimisation( ie. to change the wavelength of detection) to get a good look at impurities in case of RS is understood. but in case of dissolution .......? I dont think for the same reason wavelength would be changed when one goes from HPLC-UV to UV-Visisble spectrophotometer.
Actually i hv seen one method, that for a drug product when dissolution was done via HPLC-UV detector the wavelength used to be set at 254 nm and the same analysis when done by UV-Visisble spectrophotometer the absorbance used to be measure at 298 nm with purified water as blank.
Why would this be....? Mind you i am talking abt Dissolution and not RS where one is more interseted in the parent peak irrespective of other peaks.
Shrikant/Harsha

Actually i hv seen one method, that for a drug product when dissolution was done via HPLC-UV detector the wavelength used to be set at 254 nm and the same analysis when done by UV-Visisble spectrophotometer the absorbance used to be measure at 298 nm with purified water as blank.
Why would this be....? Mind you i am talking abt Dissolution and not RS where one is more interseted in the parent peak irrespective of other peaks.
Dissolution of formulated products has to avoid any other components, so if the analysis was on a UV Spectrophotometer without chromatography, the detection wavelength may be longer, but not at the optimum. Also, some dissolution buffers may absorb at lower wavelengths.

For HPLC, the interfering comounds are separated, so a lower wavelength can be used, perhaps even the compound peak.

254 nm was the standard fixed detector wavelength because a Hg lamp was used ( Zn gave 214 nm ), so many methods used that wavelength , regardless of where the peak was. Note that the selected spectral band width can also affect results.

Bruce Hamilton
6 posts Page 1 of 1

Who is online

In total there are 26 users online :: 3 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: ahem2, Google [Bot], Semrush [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry