PJ8 » Wed Nov 22, 2006 4:45 pm
Ideally develop a method within the pH range of the column. There's plenty of high pH stable columns (google them, I don't want to suggest some and get accused of manufacturer bias!

)
If you
must have sperisorb silica and
must have a pH outside it's stable range then you should ensure you rinse high pH mobile phase out of the column after each usage with whatever storage solvent is recommended by the manufacturer (probably MeCN) to ensure it doesn't sit and disolve for extended periods.
Edit: Of course don't wash with neat organic if using a buffer which will crash out in your lines in neat organic. Wash with 1:1 organic:Aqueous or with water then organic.