“ERR PUMP HOME” problem in autosampler shimadzu prominence S

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hi every one. I was working on Shimadzu LC connects with ABSiex 4000. MY last night batch was interrupted to stop and there was an error message showed at the front panel: err pump home. I did try power cycling and cleaning the pump glass rod but there was not improvement. Any suggestion is highly appreciated.
Check the optical sensor that helps the stepper motor determine its position. Probably a little fuzz needs to be blown out.
Thanks,
DR
Image
Hi DR,

Thanks for your advises. Would you mind to give me more specific instruction, like where those optical sensors are and how I check them?

Thanks;
Depending on the pump model , there is either reflective sensor , or optocoupler type sensor.

On reflective sensors , emitter and detector are side by side and there is a shiny surface at the motor cam, so the control circuit can detect the home position by the light reflected.

On optocouplers , there is a slot , emitter on one side ,and the detector at the other side and between there is the motor cam.The control circuit detects the home position by the light passing through the cam slot.

On both types ,if the light reflecting or passing is weak ,because of dirt accumulation , the sensor can not send the necessary voltage levels to the control circuit and an error state displayed.

Of course ,there are other possibilities , such as , broken sensor , faulty control board , faulty voltage levels etc , but first clean the sensor.

In order to reach the sensor , remove the cover of the pump , and locate the pump head . Also remove the pump head cover , the sensor located near the pump motor cam.
Hi Usman,

Thank you very much for your instruction and seems it is challenge to me. I will try to check and see whether I can fix it.

Thanks again.

NYMM11
5 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry