I had 4 decades experience with OTC products and cGMP.
I'm to agree with Tom Jupille's remarks above, no need to do a re-validation. I'd just document everything in lab notebooks or similar, and explain why stuff was done.
Of course, my pointy-haired boss would go opposite to that, as he always chose opposite to my opinions.
cGMP, USP, ICH - none specify a brand or type of instrumentation. Everything is up for interpretation by an FDA inspector, whether he/she got lucky, had a bad day, etc.
For example: cGMP guidelines say an instrument must be requalified if moved, but what constitutes a move? Is 6 inches down the bench, to an opposite lab bench, etc., a move???
Does 280nm using a UV detector mean 280nm on a DAD detector is not equivalent? Is 0.5ml/min from a quaternary pump different than 0.5ml/min from a binary pump? What if an isocratic method used pre-mixed solvent - is that different than having the pump mix them? What if the mixing valve on a quat pump was set to 100% as opposed to bypassing the mixing valve? It's pretty much impossible to specify EVERYTHING, one could never get any work done.
USP <621> even details what changes can be made in USP procedures without re-validation, and FDA ORA 4.5 even broadens that to in-house validated methods. I read those documents from the printed word, but pointy-haired boss liked to be devil's advocate 100% of the time (that baseline isn't EXACTLY flat, that calibration curve looks a little quadratic, etc.).