Hi all,

At May I encounter nitrogen bias high, then found out tat dftba vial is empty. Then refill, it look normal after that. However, after one month ltr, nitrogen become 44%. The interface is tighten well. May I know the root cause

Thanks
Seet
What about oxygen?
... May I know the root cause

Thanks
Seet
A leak is the culprit unless you messed up with carrier gas (cylinder, trap).
Did it appear when a tank was changed?
Helium has been known to have nitrogen in it. Even though you paid for it to be pure.
It could also be a leak between the tank and scrubber if you don't see oxygen.
I'd plug the flow to the MSD and run the tune, see if that reduced the nitrogen. If so, that would isolate the issue to the GC; if not, to the MSD.
5 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry