Hi everyone,

Is this fuel gas filter finished?
[url]fuel gas filter finished[/url]

Could this be the reason for my second post?

[p.s. you guys are awesome!]
Hello,

https://imgur.com/a/KjVUr

I currently run the above conditions for over 6 months, no problem.

I come back after scheduled maintenance from JayTee and the front inlet gas no longer reaches ready state.

When I decrease the make up flow, ready state is active and I obtain a spectrum which has higher adsorption than my calibrated sample.
I also noticed that the column flow decreases from 5ml/min gradually and the Electrometer increases from 42pA.

Additional information, another two GC's work from the same hydrogen generator.
What type of hydogren generator are you using? Assuming you use the generator for carrier and fuel, you're at ~1200 mL/min for that single GC. Add two more, and you may be exceeding the capacity of hydrogen generator. When the GC is in that state, what's load on the generator? Many will be able to display this.
What type of hydogren generator are you using? Assuming you use the generator for carrier and fuel, you're at ~1200 mL/min for that single GC. Add two more, and you may be exceeding the capacity of hydrogen generator. When the GC is in that state, what's load on the generator? Many will be able to display this.
Hi Jake,
the Hydrogen Generator is a H2PEM-100 / H2PEM-165 / H2PEM-260 / H2PEM-510.
We have a separate nitrogen generator on another line.
We have compressed air coming from the department.
I will try a display the load on the generator when running today.
Yes, we had spares
"H2PEM-100 / H2PEM-165 / H2PEM-260 / H2PEM-510" Which one do you have? That's a line of hydrogen generators by Parker. The number represents the maximum flow capacity in mL of each model. So if you have an H2PEM-510, its maximum flow is 510 mL/min. The pictures you posted show a total hydrogen flow of 1100 mL/min, which far exceeds the capacity of the hydrogen generator.
6 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry