taurine in energy drinks

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
hello, is there anyone who can tell me something about determination of taurine in energy drinks? i have a method using hplc uv-detection (after derivatisation with dansylchloride), which works perfectly for analysing sugar free drinks and components, but when it comes to a sample with a 'normal' sugar concentration, it is not possible to detect the right amount of taurine in this sample. i have no clue why this is so. can you help me, please :?

Dansyl chloride will react with amines and alcohols. Reaction with amine is much faster but considering that you have sugars in higher concentration with multiple hydroxyl groups and that you amine is zwitterionic you are equalizing these two reactions (between sugar and dansyl chloride and between sugar and taurine). If you are not using enough dansyl chloride some of your taurine is not derivatizing and you have lower reading. Also derivatized sugars probably causing interference. You need to use excess darivatizing agent in order to address these two issues. Also you might need method to remove derivatized sugars from derivatized taurine.

Another approach would be to use different mechanism to separate sugars from taurine and different detection technique (ELSD, LC/MS)

Or use an amine-selective derivitization reagent. The much-maligned OPA/thiol chemistry does not react with alcohols. Fluorescamine is also amine selective.
Mark Tracy
Senior Chemist
Dionex Corp.
wow, didn't expect such a reply so fast... thank you for your tips, i used an other dilution (100 instead of 10)of my sample and now it works...


marinad

marinad,
would you mind kindly explaining to us what's going on? why 10 dilution not work and 100 works?
Excel

By diluting the sample 100x versus 10x, but keeping the Dansyl chloride the same, the amount of species competing for the reagent goes down 10x. Since there is sensitivity to burn with fluorescence detection, and lots of taurine in the sample, this works. The original problem was that the sample had too many substances competing for the reagent, and the taurine was not being completely converted before the reagent was exhausted. This is a general problem for pre-column derivitization methods, and must be addressed as part of the development process.
Mark Tracy
Senior Chemist
Dionex Corp.

Mark,thank you.
Jim
Excel
7 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry