Our company just purchased a line of OTC products to go with ours. We have had similar OTC products, with same actives, for 25 years, and have always used the USP procedure on ours. Now, with the acquisition of the new brand names, our QA department insists that the USP assays are for the active ingredients themselves, and cannot be used as is for assay of the finished products unless the USP procedure is validated for the finished products themselves. Previous to this (and per my department SOP), USP and other compendial methods were taken as not needing validation, just needing a verification that the procedure works in the products. Does anyone out there really know, or can offer some expert advice? I think our QA may be trying to exert their "power", or is tremendously afraid of putting a document in our cGMP file and having to "defend" the science.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Presumably, you have been inspected / audited by the FDA at some point during those 25 years. If that SOP was in place, and if it passed muster with the FDA, then "if it ain't broke, don't fix it".
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 2
- Joined: Sat Oct 14, 2006 4:17 am
Mr. Jupille: our company's products of this type have been manufactured at co-packers for at least a dozen years, and either no FDA inspector ever questioned their assay procedure, or was OK with it, but there is no way to know. As to the newly-acquired products: yes, their (former) facilities were inspected by the FDA numerous times, and there was never an issue about this (and, apparently, this is considered a "universal" procedure for this active). But my company's QA department is "special". They didn't have an SOP in place detailing that they were just using the USP procedure either, or their rational for using that procedure.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Strictly speaking, your QA people are correct. There is no guarantee that that there isn't something in your formulation that would interfere with / invalidate the USP procedure. That said, I find it hard to visualize that these products have been made and analyzed for over a decade without having an SOP in place on how the analysis is carried out.
Someone needs to go over the documentation history on this stuff before a "feco-ventillary collision" occurs.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
You will have to revalidate.
As mentioned, you are dealing with a new formulation so you have to revisit (at least) all aspects of the validation that deal with specificity and accuracy. You should also repeat your forced deg. studies as you may not have the same impurity profile as with the legacy product. If the new formulation is pretty much the same as what you had adn the sample prep. is about the same, you may be able to skirt some robustness work.
As mentioned, you are dealing with a new formulation so you have to revisit (at least) all aspects of the validation that deal with specificity and accuracy. You should also repeat your forced deg. studies as you may not have the same impurity profile as with the legacy product. If the new formulation is pretty much the same as what you had adn the sample prep. is about the same, you may be able to skirt some robustness work.
Thanks,
DR

DR

-
- Posts: 193
- Joined: Thu May 11, 2006 11:10 pm
Dear all:
I agree with DR and Tom. USP monographs are mostly for pure materials, and the assays (if exist) are for pure materials. Monographs for compounded products with active ingredient(s) are rare!
If you use the assay method directly from USP for your products, the matrix may interfere with the assay, and alter the results. We have had many problems with USP methods that did not work for our products. We had to develop the methods internally (of course, we do the validation!).
As quoted by anonymous, "the USP assays are for the active ingredients themselves, and cannot be used as is for assay of the finished products unless the USP procedure is validated for the finished products themselves," I agree completely with your QA's position. Perhaps you must enhance compliance, by validating USP methods on your finished products. It is good practice to check system suit, and accuracy.
Alfred.
I agree with DR and Tom. USP monographs are mostly for pure materials, and the assays (if exist) are for pure materials. Monographs for compounded products with active ingredient(s) are rare!
If you use the assay method directly from USP for your products, the matrix may interfere with the assay, and alter the results. We have had many problems with USP methods that did not work for our products. We had to develop the methods internally (of course, we do the validation!).
As quoted by anonymous, "the USP assays are for the active ingredients themselves, and cannot be used as is for assay of the finished products unless the USP procedure is validated for the finished products themselves," I agree completely with your QA's position. Perhaps you must enhance compliance, by validating USP methods on your finished products. It is good practice to check system suit, and accuracy.
Alfred.
-
- Posts: 231
- Joined: Tue Feb 15, 2005 9:51 am
It seems to me that sometimes when it comes to validation, commonsense goes out the window - it has taken us over a year to validate a CDS - Empower!
If we took the same view when buying a car it would stand in the drive for a six months before it was let out on the road!
If we took the same view when buying a car it would stand in the drive for a six months before it was let out on the road!
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Is this because you're validating all of your custom fields & calculations before you use it? [/hijack]- it has taken us over a year to validate a CDS - Empower!
Thanks,
DR

DR

-
- Posts: 231
- Joined: Tue Feb 15, 2005 9:51 am
I wasn't involved in the validation so I can't answer that question. My point was the validation industry seems to have taken on a life of its own. I the past I have impemented a CDS system - we boughtit put on the bench read the manual and started using it - took a week or two. If control samples were OK we assumed the system was OK especially as it was being used by 1000s of others. Will laboratories be validating Vista when it arrives?
9 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: John Guajardo and 23 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: John Guajardo and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science